I've received a loaner probe from Bruker while ours is being repaired but I'm not getting any timely response on instructions on calibrating it for use. Anyone have a resource or documentation? Thanks, Ken asked Jun 15 '12 at 12:58 Javaslinger |
If it is the same type of probe that was replaced, then you only need to tune and match each channel to use it. The pulse widths and power levels from the original probe should be close enough. answered Jun 18 '12 at 10:07 |
Ok, I take back my negative post regarding Bruker tech support. In the end they were tremendously helpful and walked me right through it. answered Jun 21 '12 at 20:48 Javaslinger @Javaslinger - the posts can be edited so you can remove whatever content you dont' like. Thanks. - Evgeny Fadeev (Jun 21 '12 at 21:03) |