Hello, I'm measuring T2 in paramagnetic materials. These values are very short (0.5-5 ms), and I have to use very short interpulse delays. On Bruker machines standard CPMG sequence looks like this: d1 p1(90°) [d20 p2(180°) d20] the number of 180° trains [number of times the sequence in square brackets is repeated] is specified in vc list. I use 12 points, from 8 to 96 pulse trains. Suppose my d20 is 40 us and my p2 is 20 us. My questions are:
Thanks! asked Jan 18 '15 at 00:26 A K |
(2) if you have an oscilloscope, you can monitor the reflection which tells you whether arcing occurs. From my experience, if you are using an liquid probe that usually runs with a power 50W, 2ms pulse should be fine. You can also make you d1 longer for better heat dissipation. I have tested CPMG with very high power, 300W, 2us pulse+10us spacing (equivalent to your d20 5us), 1k pulses. It still works fine. answered Feb 27 '15 at 12:48 |