Hello, posting this on behalf of my friend.

The goal is to set up a gaussian shaped pulse so that bandwidth is X Hertz. How to select duration and power of the pulse to satisfy that that?

Thanks!

asked Apr 01 '10 at 15:30

Evgeny%20Fadeev's gravatar image

Evgeny Fadeev
5771

updated Apr 01 '10 at 16:58


One Answer:
i like this answer (click again to cancel)
1
i dont like this answer (click again to cancel)

The answer depends on which Gaussian pulse is in use. There is a simple gaussian (for which I give hopefully the real answer below), there are also G3 (consists of three different gaussian pulses back-to back and is used for inversion) and G4 (four-pulse cascade used for the excitation) pulses. Cascade pulses have excitation or inversion profiles closer to rectangular.

Gaussian pulse can also be used for selective saturation (2,4) (which is applied for detection of binding of small molecules to large bio-molecules in STD experiments) , where a train of pulses separated by a short delay is applied for some period of time. Here selectivity will depend on the total duration of saturation time and parameters of the pulse, but it should be easy to run an STD-type of experiment with a small molecule alone first to make sure that saturation does not affect it.

Setting up single gaussian pulse

Here is the link to the original paper for a single gaussian pulse. A 10ms gaussian pulse calibrated for 90 degree on-resonance flip angle will introduce almost no excitation at offsets >300 Hz.

So if you apply a single gaussian pulse like that for selective excitation the closest other resonance must be farther than 300 Hz away. If you need better selectivity - then just proportinately increase duration of the pulse and re-calibrate it again so that it has 90 degree flip.

References

  1. Christopher Bauer, Ray Freeman, Tom Frenkiel, James Keeler and A. J. Shaka Journal of Magnetic Resonance (1984) 58(3), 442-457
  2. Moriz Mayer, Bernd Meyer. J. Am. Chem. Soc. 2001, 123, 6108-6117
  3. Lyndon Emsley and Geoffrey Bodenhausen. Chem. Phys. Lett. 165(6), 469-476.
  4. Saturation transfer difference (wiki page)
  5. Signal saturation in NMR (wiki page)
link

answered Apr 02 '10 at 06:54

Evgeny%20Fadeev's gravatar image

Evgeny Fadeev
5771

updated Apr 02 '10 at 13:15

Your answer
Please start posting your answer anonymously - your answer will be saved within the current session and published after you log in or create a new account. Please try to give a good answer, for discussions, please use comments and please do remember to vote (login to vote)
toggle preview

powered by CNPROG