I had just finished refilling the nitrogen and powering up the NMR after the holiday shutdown and went through the normal startup routine (cf, ii, cfbsms). When I ran lockdisp I could not get the lockdisplay window to appear, appearing to be stuck on 'checking rs-channel' and failing after 11 attempts. Any help in the matter would be appreciated. System details: XwinNMR 2.7 on SGI O2 with IRIX. Bruker Avance DPX400. Things checked already:
asked Jan 02 '13 at 06:37 Norman Chu |
Problem resolved - fix from Bruker: On the LCB board of the BSMS, while the console is powered and connected to the computer, disconnect, and then reconnect the RS232 cable from the LCB board, then reconnect it, and run lockdisp again. - Norman Chu (Jan 02 '13 at 10:54)