Hi all:) I apologize for my simple ( or well, stupid) question: I was wondering if it is possible to run a 2D 13C19F experiment on a Bruker Avance spectrometer with BBFO probe. I was under the impression that 19F is run on the 13C channel, so its not possible.. something like this. Is this correct? Thanks!!! |
Hello, maybe you have luck, ... please look into the specifications of your Bruker probe! There are two different types of it: BBOF and BBOF-Plus.
Yours sincerely, Uli Haunz |
There's only chance for Bruker BBFO probe to tune H coil to 19F if you see the labels on RF connectors at the probe body: 2H | HF | BBF - there's a rear modification of BBFO probe (but BBFO plus is not the same, it's just BBFO with extra sensitivity) that gives you a way to measure 13C-19F correlations If the labels are the following: 2H | H | BBF - there's no chance to tune 1H coil to 19F (for ATM probes at least) |
Hi, yes, as far as I can see it, such an experiment is not possible on a BBFO probe. |
Hi, Yeah it doesn't seem like 13C{19F} can be done with a BBFO 2 channel probe, for the reason you stated. |
Hi, Yeah it doesn't seem like 13C{19F} can be done with a BBFO 2 channel probe. - RudrakshLethbridge (Nov 16 '12 at 09:29)