Why do we trun off sample spinning when we do NOESY experiment?

asked May 25 '12 at 06:53

Kiu%20Chor%20Sham's gravatar image

Kiu Chor Sham
11

updated Jun 03 '12 at 17:23

Evgeny%20Fadeev's gravatar image

Evgeny Fadeev
5771


2 Answers:
i like this answer (click again to cancel)
3
i dont like this answer (click again to cancel)

Spinning your sample is not reccomended during 2D experiments, especially if you are using gradient-enhanced techniques. In a 2D NOE experiment, spinning the sample wil increase the rate of diffusion in your sample and some of the NOE might be lost as the molecules of interest bounce around in your sample. Remember that NOE is transferred through space and it might even affect neighbouring molecules if they collide, transferring some nuclear magnetization. You don't loose sensitivity by spinning but you don't emprove lineshape either. If you are looking for better lineshapes try shimming better and using a proton observe probe (inverse probes on Bruker spectrometers: BBI, SEI, etc). Spinning will result in a loss of NOE in your sample and your spectrum might not be useful anymore, especially if your sample is dilute.

link

answered May 30 '12 at 23:12

Chisbora%20Cristian's gravatar image

Chisbora Cristian
51

The diffusion or NOE themselves cannot be affected by spinning, but physical movement of parts of the sample will be, especially in the transverse plane> So the spatial magnetization distribution encoded by the gradients (esp. by the X and Y gradients) will be hard to decode with the spinning on. - Evgeny Fadeev (Jun 03 '12 at 17:22)

i like this answer (click again to cancel)
3
i dont like this answer (click again to cancel)

Spinning should probably be turned off for any 2D (or higher) experiment. Spinning will cause a modulation of the signal as a function of the evolution time, which will result in T1 noise (big streaks in the F1 direction of the transformed data). The shim on modern instruments is usually so good that it is rarely necessary to spin 5mm samples, even for 1D.