Hello, why choloroform gives a negative triplet instead of a positive(just one hydrogen on the carbon) in DEPT 135. Can you give me an explanation

asked Jul 25 '14 at 05:15

zikoooo's gravatar image

zikoooo
3

Actually, I found this signal 1;1;1 in all my spectra. I am using a Bruker 400Mhz. Which parameters should I change to get a "normal" positive signal for the chloroform which is a impurity in the deuterated solvent(chloroform). Cheers, - zikoooo (Jul 28 '14 at 01:48)


3 Answers:
i like this answer (click again to cancel)
1
i dont like this answer (click again to cancel) the author of the question has selected this answer as correct

From what you have described, it sounds like you have run a DEPTQ experiment. Check the pulse sequence for your experiments (typical Bruker pulse sequences for DEPTQ are deptqsp, deptqgpsp, and deptqgpsp.2). Glenn Facey has done a comparison of DEPT-135 and DEPTQ experiments on his blog (http://u-of-o-nmr-facility.blogspot.com.au/2007/11/dept-and-deptq.html).

link

answered Jul 28 '14 at 16:38

cyg's gravatar image

cyg
36

Thanks, the link is really helpfull. - zikoooo (Jul 29 '14 at 01:20)

i like this answer (click again to cancel)
1
i dont like this answer (click again to cancel)

Is it deuterated chloroform? In that case it should give a 1:1:1 intensity triplet in 13C spectrum, and shouldn't give any signal in dept. There is probably a misadjustment of pulse length, or you didn't do a multiple of 32 scans for phase cycling in your dept expt.

link

answered Jul 27 '14 at 22:56

Anne%20Baudouin's gravatar image

Anne Baudouin
36

Actually, I found this signal 1;1;1 in all my spectra. I am using a Bruker 400Mhz. Which parameters should I change to get a "normal" positive signal for the chloroform which is a impurity in the deuterated solvent(chloroform). Cheers, - zikoooo (Jul 28 '14 at 06:16)

I am doing 256 scans each. - zikoooo (Jul 28 '14 at 09:14)

i like this answer (click again to cancel)
1
i dont like this answer (click again to cancel)

Perhaps it is not chloroform. If you see a triplet with 1:2:1 intensities, that would suggest there is a CH2 group, which yields negative amplitude in DEPT 135.

If the intensities are 1:1:1, that would be deutero-chloroform. Ideally it should not appear at all (no H, therefore no H-C transfer). If it does appear, it may be because the recycle delay is too short (first scan gives larger signal which defeats the phase cycling). Try a longer recycle delay, or use some dummy scans, and it should disappear.

Note added soon after posting: I see that Anne Baudouin responded similarly while I was typing.

link

answered Jul 27 '14 at 23:06

Tony%20Bielecki's gravatar image

Tony Bielecki
131

updated Jul 27 '14 at 23:11

Actually, I found this signal 1;1;1 in all my spectra. I am using a Bruker 400Mhz. Which parameters should I change to get a "normal" positive signal for the chloroform which is a impurity in the deuterated solvent(chloroform). Cheers, - zikoooo (Jul 28 '14 at 06:16)