i like this post (click again to cancel)
0
i dont like this post (click again to cancel) remove favorite mark from this question (click again to restore mark)

HI I am faceing spining problem on av-300 bruker instrument. I have change spinner,clean shim stack but the problem is still there have any body idea what i have do now,I have check by turning on and off spinner change spin rate but it not getting spinning. IS it electronic fault?

with regard feroon

asked Sep 10 '11 at 11:14

feroon's gravatar image

feroon
21


One Answer:
i like this answer (click again to cancel)
0
i dont like this answer (click again to cancel)

Check spin airflow by disconnecting tubing at upper barrel and turning spin on/off. Mark the spinner so you can see if spinning by looking into the upper barrel with a light. If spinning, check tachometer wiring and clean tach optical sensor, or reduce VT airflow.

link

answered Sep 12 '11 at 09:31

Jerry%20Hirschinger's gravatar image

Jerry Hirschinger
199

Your answer
Please start posting your answer anonymously - your answer will be saved within the current session and published after you log in or create a new account. Please try to give a good answer, for discussions, please use comments and please do remember to vote (login to vote)
toggle preview

Tags:

×4

Asked: Sep 10 '11 at 11:14

Seen: 3,606 times

Last updated: Sep 12 '11 at 09:31

powered by CNPROG