I have a problem with gradshim (gradshimau) in topspin. The problem is that im getting spitting on solvent peak (and all other peaks as well) due to poor Z-Shimming.

I have tried 2 ways of creating new shimmaps:

1 I used 3%CHCl3 in Aceton d6 (but I'm not sure this is the best sample to use due to the long relaxation time of the sample). Anyways, i created a new GSHIM file and new shimmapping. When runing Gradshimau Users experienced that they need to manualy shim (especially) Z2 due to the splitting of the peaks.

2 The same procedure as above but with a sample that contained 90% D20 / 10% H20 + CuSO4, im attaching the shimmap that i created from this sample. the problem is stil that we end up with unwanted splitting due to poor Z2 shiming.

alt text

Any sugestions on how to continue?

asked Oct 27 '10 at 06:22

jonis's gravatar image

jonis
11

updated Oct 27 '10 at 06:31

hey jonis, maybe you can come up with a better title for your question, so it sounds like one? it usually makes more people take a look. - Evgeny Fadeev (Oct 27 '10 at 11:33)

Your shim map looks a bit noisy. You might want to look at your parameters (pulse lengths, number of scans, echo times, etc.) to make sure if things are set up correctly. Also look at you shim iteration groups and the length that is used for the shim. If this is for TopSpin 2 or up, use TopShim. - Kirk Marat (Oct 28 '10 at 12:25)


3 Answers:
i like this answer (click again to cancel)
2
i dont like this answer (click again to cancel)

Hi, according to Varian grad shimming manual (maybe Bruker has a manual like that too?) chloroform in deuteroacetone is used for shimming for the best lineshape. The recommended recycling delay is 6-12 seconds and the steps are:

  • shim on z1, z2, x1, y1 manually
  • set up the grad shimming pulse sequence
  • record the shim map
  • gradient shim on Z shim set
  • shim x1, y1, xy and zy (do not touch Z shims)
  • repeat gradient shimming

Perhaps a couple of first times you'll need to shim higher order x and y's. But you need to know which one's to adjust on your magnet (for example on our magnet I've been advised not to change X3 and Y3). I occasionally adjust xy, x2-y2 and third order x and y's except x3 and y3.

Take a look at this shimming manual by Acorn NMR, it explains typical interactions of shims.

link

answered Oct 27 '10 at 11:31

Evgeny%20Fadeev's gravatar image

Evgeny Fadeev
5771

i like this answer (click again to cancel)
1
i dont like this answer (click again to cancel)

If this is for small molecule work, you could try spinning your sample at about 4 Hz. This might get your line widths down.

link

answered Oct 27 '10 at 11:50

Scott%20Robson's gravatar image

Scott Robson
156

updated Oct 27 '10 at 11:51

spinning averages out imperfections in transverse shims, they might as well be fixed by adjusting X & Y's - Evgeny Fadeev (Oct 27 '10 at 12:04)

true, but if it can't be done, spinning might help. If the problem is with Z shims the solution may very well be complex. - Scott Robson (Oct 28 '10 at 13:44)

i like this answer (click again to cancel)
1
i dont like this answer (click again to cancel)

There are two problems with using the standard lineshape sample for mapping: long 2H T1, and convection. The lineshape sample is in acetone, which has very low viscosity and boiling point, so convection is typically a big problem. In the gradient shimming experiment, two gradient echo images are acquired, with short and long echo times, and convection can significantly reduce the signal in the long echo experiment. If the spins move in the echo delay, they experience different effective gradient strengths in the encoding and decoding steps and the signal is thus not refocused. For that reason, DMSO or D2O samples are normally preferred for mapping - convection is minimised, but you still have a strong 2H signal.

As Kirk indicated, your shim maps seem noisy. This could come from poor parameter settings - what changes have you made from the standard gradshim1d2h parameter set? The most basic things you need to optimise are the 2H pulse length, and the number of scans. Note that the 2H 90 degree pulse needs to be in the prosol table, as since xwinnmr 3.5 gradshim has used getprosol for the 2H 90 degree pulse, and the pulse in the parameter set will be ignored. The pulse program then uses p1*0.15 for the excitation pulse. You probably want a minimum of 32 scans for reasonable S:N, more if you are at low field (300 MHz or less). Note that the D1 used is very short (consistent with the small flip angle excitation pulse), so 64 scans might take e.g. 15 seconds per echo measurement. You can find more details about parameter settings in the guide that was published in the Bruker Spin Report - first Google hit if you search for: gradshim manual spinreport.

For iteration files, larger window sizes should improve the hump, but if the window is too large the effect of the noisy points at the ends of the shim map will distort things. Does a window of 15 work with the above maps? Normally you want a larger window, as this will optimise the hump, but not so large that the resulting half width becomes broad, but running with a smaller window may indicate if the problem is with the whole maps, or just that the edges have too low S:N.

Finally, if the starting shims are not good, your maps will not be either. If you are doing the mapping without spinning, the off axis shims need to be reasonably good as well as the Z-shims, so adjust x, y, xz, yz by hand if you are going to run nonspinning. The usual advice is that you should run the mapping as you will do the shimming, i.e. if you run all your samples spinning, generate the maps whilst spinning too.

link

answered Oct 30 '10 at 08:34

Pete%20Gierth's gravatar image

Pete Gierth
401