Is it possible to use NMR for limit testing of trace impurities in pharmaceutical drug substances? What are the considerations? asked Apr 26 '10 at 10:32 |
With oversampling and DSP dynamic range should not be much of a problem these days. It is now up to 22 bits, which would be a dynamic range of ca 2 million to one. answered May 06 '10 at 07:25 Kirk Marat |
The following references might be relevant.
They write "...Detecting a low level-level (<10ug/ml) small molecule impurity in the presence of active protein and excipients at much higher concentrations is a daunting analytical problem. In many cases, NMR is an excellent method for solving such challenging analytical problems..." I do agree with Kirk Marat that, in most cases, other techniques would be preferable. answered May 07 '10 at 05:47 sekhar Talluri |
You definitely can use an internal standard at specific concentration and use ratios of peak integrals to measure concentrations by NMR, but one thing to be mindful about is distortion of peak intensity by saturation - i.e. pulsing too frequently and thereby not allowing for sufficient relaxation of magnetization between scans. A book "High-Resolution NMR Techniques in Organic Chemistry" (ISBN 978-0-08-042798-0) by Timothy Claridge has a very good explanation of this experimental detail in Chapter 4 So - yes it must be possible to set upper limit of impurity content using NMR data, the lowest number that can be reported this way depends on sensitivity of the instrument and amount of sample in the tube or rotor (for solids). I hope that somebody will correct me if I'm wrong. answered Apr 28 '10 at 14:03 Evgeny Fadeev Thanks for the reply! - Chris Kirchner (Apr 30 '10 at 15:29) |
Hello Chris, could you explain what limit testing means? - Evgeny Fadeev (Apr 26 '10 at 10:39)
A limit test is when you want to test whether something is above or below a certain level. For qNMR, i've been told there is a dynamic range restriction that does not allow quantitation (at least linear?) under a certain level for the analysis of trace impurities. - Chris Kirchner (Apr 28 '10 at 10:55)
However, if we can get enough sensitivity and can detect/see the impurity at a given level, can we use this level of detection as a limit and say whether the impurity in our actual sample is above or below this limit? - Chris Kirchner (Apr 28 '10 at 10:56)
I guess this sounds pretty comensensical, however, i was shocked when i couldn't find one single article on limit testing by NMR so just being comprehensive... - Chris Kirchner (Apr 28 '10 at 10:56)