We have got a donated AQR/p with Spectrospin 300 magnet and there is no helium inside. It was sitting in the warehouse for a long time with no helium inside. There is a possiblity of dried helium or quenched magnet. I need help to now how to recover magnet from this condition and get it up and running. Already thinking about borrowing a magnet power supply (0-10V, 0-100A) but I need some manuals to tell me how to use this power supply and with what ramp increase the current and the time and procedure to switch to persistent mode. you can also email me directly on agooseventapegie@gmail.com

thx

asked Sep 19 '13 at 04:39

agoo's gravatar image

agoo
1


4 Answers:
i like this answer (click again to cancel)
3
i dont like this answer (click again to cancel)

The procedure to recover a quenched magnet is extensive, and requires much more than just a single power supply. Just to test it for viability, you need an energisation lead, shorting plug removal tool, and the magnet manual specific for that magnet serial number. If the magnet passes room temperature resistance tests, you then need a gantry, hoist, magnet seals, vacuum pump, leak tester, blowout tube, liquid helium transfer line, shim power supply, SC switch heater power supply, and many other accessory pieces of specialised cable and tubing in addition to a magnet power supply. This is not a job for you to succeed without hiring a professional magnet service engineer, which will cost much greater than $10,000.

You also need to test the AQR to see if it operates. That means testing each power supply and cooling fan, the host computer and software, then the remainder of the electronics.

"Always look a gift horse in the mouth!" (You might find no useful teeth.)

link

answered Sep 20 '13 at 08:08

Jerry%20Hirschinger's gravatar image

Jerry Hirschinger
199

I totally agree with Jerry. This is NOT something that you should try yourself without the proper training and equipment. Even cooling the magnet is a complex precedure. Any magnet that has quenched should probably be re-pumped and, depending on age, have the seals replaced. - Kirk Marat (Sep 25 '13 at 09:30)

i like this answer (click again to cancel)
1
i dont like this answer (click again to cancel)

Replacing magnet seals and that sort of thing would be a good idea too. We recently did this (with the help of the Bruker engineer on-site), and it was a major job. Not one to do without the engineer and the correct power supply (which is not a normal standard thing, but a big monster with huge cables that weighs a lot). Your magnet needs the correct charging rod (there are quite a few different ones), and you will probably also need fittings to pump it down again (unless it is a really really old (pre-1990-something) one where the fittings were built into the magnet.

At the helium price around here the quenches we didn't have during recommissioning and the helium we will save because the magnet is optimally put together pretty much covered the cost of the engineer.

If it was decommissioned properly then in the magnet manual there should be a discharge record. There should also be a charge record, from when it was first charged, which is helpful for recharging. If you can find the manual it would tell you if it was discharged correctly - which might also help you work out if it is worth resurrecting.

Kate

link

answered Sep 27 '13 at 23:11

Kate%20Nairn's gravatar image

Kate Nairn
11

i like this answer (click again to cancel)
1
i dont like this answer (click again to cancel)

The primary knowledge you need is a copy of the magnet manual for that specific coil. A Bruker magnet manual will be titled “SC-MAGNETSYSTEM”, and the coil number is on the label on the side of the magnet. You may contact Bruker to receive a copy of the manual if it is not with the system. Bruker NMR magnets are not high-current. A 650 ampere power supply is not necessary. 100 amperes is likely more than sufficient. There are also 3 power supplies needed, not just one. The internal structure important for electrical connections and for vacuum connections , as well as the power supply requirements are detailed in the manual, and the precise currents needed to approximately reach the field homogeneity necessary to do NMR are also detailed in the manual for that coil number. Even before filling with liquid helium, there is an internal siphon connection which must be reached to remove contamination from the helium chamber. This is why you need a magnet manual – to know what pieces and parameters are required.

The most import piece of equipment is the energisation lead (also called a stick or rod). There is NOT a detailed diagram of the stick dimensions in the manual. The stick contains superconducting wire, and it is not possible to measure the mating connector, so a stick is not practical to fabricate. If this item is not in hand with the magnet, it simply cannot be operated. Before you seek a power supply, you need a Bruker magnet energisation stick and the magnet manual. Then you will be able to test the magnet and shim resistances at room temperature to know if it might be damaged electrically.

Without newly installed cryostat vacuum seals, the magnet will require an excess amount of cryogens to maintain operation. One cannot operate an NMR magnet cryostat while attached to a vacuum pump, because the liquid helium evaporated gas travels backward through the vacuum pump by means of cryopumping. Evaporated helium gas also enters easily through even the smallest of seal leaks, soon making the cryostat impossible to operate due to helium gas contamination of the vacuum insulation. When helium gas contamination reaches the superinsulation layers within the cryostat, it cannot be removed and spoils the vacuum insulation of the cryostat forever. Before attempting energisation, I recommend to install new vacuum seals and test thoroughly for leaks with a helium leak detector.

In addition, you must have reliable delivery and funding to buy cryogens. Approximately 50 liters of Liquid Nitrogen are required every week. Approximately 100 liters of Liquid Helium are required every 6 months. In addition, several hundred liters of both LN2 and LHe may be needed for each quench or energisation.

Only until you have a reliable cryogen supply, the magnet manual, an energisation stick, and a reliable vacuum seal, can you hope to operate a magnet power supply.

link

answered Oct 04 '13 at 09:58

Jerry%20Hirschinger's gravatar image

Jerry Hirschinger
199

i like this answer (click again to cancel)
0
i dont like this answer (click again to cancel)

Jerry

Sorry my fault not telling that this is a Bruker AQR/P Spectrospin 300 complete previously working NMR machine. Just ran out of the helium. They had to turn the magnet off (hopefully according to the proper procedure).

Putting brieflly what I have understood so far is after filling N2 and He I should:

1- Connect the power supply to the magnet ports on the top of Spectrospin tank 2- Turn on the magnet persistent switch heater 3- Program the powersupply for about 0.2A / min and turn it on 4- After magnet current reached the maximum set current (about 100A) turn off persisten switch heater 5- Turn down power supply current to zero gradually and turn it off at zero 6- disconnect power supply

I need more detail explanation on these items and in case if I have missed anything here? A video showing the procedure could be lot better.

link

answered Sep 20 '13 at 09:00

agoo's gravatar image

agoo
1

updated Sep 20 '13 at 09:18

That is a very simplified description of only a small part of the work needed. This is not a proper forum for training a complex procedure. - Jerry Hirschinger (Sep 30 '13 at 10:25)