We generally use Dual to run 13C and BBI to run 2D. After changed the probe, the command "edhead" was used to set the probe. Put the sample tube, lock the solvent, and then type "atma" to tune the probe. We always do it like this, but now we can not tune the proton after installed the BBI probe (13C is OK). The dip can not be found by "atma", and "atmm" was also not work on forming a dip. What is the most possible reason for this error? How to solve it and avoid it in the future ? Thanks. (Instrument: Bruker 400 MHz, Topspin) asked Aug 22 '11 at 22:57 |
answered Aug 28 '11 at 22:02 VVK |
Most common error after a probe change is an incorrect cable connection. Double check the cable routing to be sure each probe coil is connected to the correct preamp. - Jerry Hirschinger (Aug 29 '11 at 06:28)