Hello, I've noticed recently that our EPR spectrometer (Bruker EMX500) has a negative offset on the baseline and it becomes very large at high receiver gain settings (goes off scale). Is there a way to tweak the spectrometer or software settings to fix this issue? Thanks! asked May 06 '10 at 08:42 Evgeny Fadeev |
The Bruker EMX software corrects the baseline afyet each scan... I can think of two reasons for the offset: 1) There is a very broad line there (dirty cavity?). To check it try to run your spectrum with larger sweep range (3000-4000 G) 2) Check the phase of the MW bridge Alex answered May 12 '10 at 11:55 |