I am running BioPack on a Varian INOVA 800MHz on a newly aquired 13C15N1H sample and I have come across an issue that I was wondering if anyone could shed some light on. When it got to the CBCAcoNH, the experiment it setup to calibrate the phi7cal is set to take ~3hours! with nt=520. I have reasonable signal (my sample is 0.8mM), so why is it doing this? Does it determine a S/N value at somepoint that maybe went awry? I set up BioPack in the following manner: create new experiment, load ghn_co and set parameters:
set pw and tof as manually determined Main Menu ->Setup->Proteins->Manag Probe Files=> Update Probe Files with these Parameters Back->Calibrate->Full Thanks for any insights! |
|
Hi, If you got to the stage of You should recall You can also recall the Check the printouts (if you did them) and look at the nt value in the parameter printout. Otherwise, you could recall the data in the order it was acquired (use a shell and cd to vnmrsys/AutoTripRes. Then do " -George Gray, Varian Inc. Thanks, George. I've fixed up formatting a little. This markup uses underscores for italics. Strings with literal underscores need to be quoted with backtics. - Evgeny Fadeev (Jan 20 at 18:50) Thank you for the help! I will check it out! - cmoody (Jan 21 at 11:52) |
|
There may be other reasons, but if "hard" 13C 90 pulse is longer than probe specification - you can check this value yourself, Biopack calibration may set decoupler power to a value above the limit set in the pulse sequences. Then routine will fail with an error something like "dpwr too high, don't fry the probe". Other reasons might be trivial - your sample being too dilute, cables not connected properly or "other hardware issue" - which can of course include lot's of things. If you can confirm this failure with a reliably good 15N, 13C-labeled sample and you are confident in cabling of the instrument - you should probably ask your facility engineer to check the spectrometer and the probe. |